What is best way to set up an MHE run? I have Combi PAL, Cyle Composer, GC Solutions (Shimadzu) and the AOC-5000 "game boy" to deal with. Seems like some combination of one or some of these parts should work for doing (for example) repeat injections (5) on a and then move on to an unknown and do 5 repeat injections on that sample.

I'm close in that I stopped GC Solution from taking remote control of the AOC-5000. That allowed for the timing of incubation/agitation to coordinate better, but I still can not get AOC-5000 to do repeat injections on the same vial. My service guy recommended setting up multiple "Jobs" on the game boy, but that didn't work either because I keep getting the "memory full" message on the game boy.

HELP!
Cycle Composer should allow you to configure repeated injections as part of a sequence. Instead of setting multiple "Jobs" on the AOC-5000, you could use Cycle Composer to define a single job for performing repeat injections (for example, 5 injections per sample) by incorporating loops. Stickman Hook
2 posts Page 1 of 1

Who is online

In total there are 346 users online :: 0 registered, 0 hidden and 346 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 346 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry