The GC analytical column is used as a transfer line: it runs up through the injection port and runs through the transfer line sleeve, and connects directly to the headspace analyzer. It's a 30 x 0.32 x 5 Elite-1 column. So the HS should be providing the carrier gas (nitrogen) to the GC. The detector is an FID.
HS settings:
Needle 80C, Transfer 90C, oven 35C, carrier 10 psi
Pressureize 3 min, inject 0.5 min, withdraw 1 min, thermo (equillibrium) 5 min, GC cycle 4 min.
Vial vent is off, high pressure injection is off, column pressure set to 10 psi.
GC Settings:
Inlet gasses off
FID hydrogen/air at 45 and 450 ml/min
Inlet at 150C and oven isothermal at 150C for four minutes.
For method development I have the thermal equilibrium step on the HS set to 1 minute, since it's just propane and I don't imagine it to have much equilibrium with water. For actual calibration and sample analysis the method requires one hour.
My problem is that sometimes I get a real nice propane peak, and sometimes I get basically nothing - just some very small baseline fluctuations. This is my first time using PE equipment and doing headspace, so I'm looking for any suggestions on whether my method settings look reasonable. This method (EPA 40e) is old and nobody seems to use it anymore so I can't find application notes.