PPC SHUTDOWN ERROR
Posted: Tue Sep 10, 2024 9:37 am
by AAK125
Hello, Could someone please tell me what to do if Clarus 580 shows the following error messages: CAR 1 PPC SHUTDOWN and SPL1 PPC SHUTDOWN. Thank you
Re: PPC SHUTDOWN ERROR
Posted: Tue Sep 10, 2024 12:52 pm
by MichaelVW
Hello
Step 1 is to check that carrier gas is getting to the inlet. Is the pressure setpoint the same as the pressure actual?
Re: PPC SHUTDOWN ERROR
Posted: Tue Sep 10, 2024 2:36 pm
by AAK125
Hello, thank you for your reply.
When I lauched my analysis, the actual pressure and setpoint pressure were same. But, midway it started beeping, and the pressure went to zero and is showed SPL1 PPC SHUTDOWN on the GC interface.
Re: PPC SHUTDOWN ERROR
Posted: Wed Sep 11, 2024 2:53 pm
by MichaelVW
If you provide detailed info, you'll probably get better answers here.
Watch the pressure during a run. The setpoint might increase along with the temperature, and maybe the supply pressure isn't high enough to keep up with it. The solution would be to make sure the supply pressure from your regulator is high enough, and (as is the case 90% of the time with GC) to fix any leaks.
Re: PPC SHUTDOWN ERROR
Posted: Tue Sep 17, 2024 7:47 am
by AAK125
Hello,
Thank you so much for your response. The problem is resolved, turns out the pressure regulator was not working properly and as a result the carrier gas pressure was not reaching the setpoint.