I've run a few other samples and they work fine, including the structurally similar cinnamaldehyde.
I don't think it's a suppression issue as a solution of 50:50 benzaldehyde:cinnamaldehyde still shows cinnamaldehyde nicely, so it's not some stabiliser in my benzaldehyde sample suppressing the signal.
Help greatly appreciated!
Conditions below.
Sample:
1 mg/mL benzaldehyde in ACN
Column:
4.6x150mm C18 5µm
Method:
Solvent A - Water (LC-MS grade) + 0.1% FA
Solvent B - ACN (LC-MS grade) + 0.1%FA
Gradient from 95A:5B to 5A:95B over ~50mins with appropriate switches etc.
System:
Agilent 1200 series HPLC
Agilent 6120 Quadrupole LC-MS
MS conditions:
ESI, 3000V
Gas temp 250°C, 12L/min
Nebuliser pressure 35psig
Quadrupole temp 100°C
We've tried a variety of fragmentor voltages and not seen much effect, perhaps some benzaldehyde fragments at higher voltages.