Thanks in advance!
EDIT:
=SOLVED=
It appears that when I was changing a rotor seal, I had to move the needle guard to get better access. When I did so, I unseated the (tiny) needle position sensor that seats at the top of the guard. I made sure the positioning hole in the ribbon was seated on its pilot peg, and the sensor was seated in its intended position. Reassembled, cycled the power, and the carriage initialized to its proper position. This resolved the issue.