Tetraethyllead non linear response
Posted: Wed Aug 14, 2024 8:11 pm
Colorado is about to require TEL measurements on suspected ground water projects. I got the 200ug/mL TEL standard 200ug/mL AccuStandard Cat# S-1263. I transfered it into 1mL Mininert vial just as I use for my 8260 standards.
I had just cleaned my source and was putting on an 8260 calibration. I got very good RSD's for my new calibration on the order of 5-7% RSD for all 79 regular compounds; except for TEL. It is distinctly non linear. I was able to fit a response factor curve (20, 40, 80, 120, 160 ppb) but any higher concentrations (240, 320, 400, 480 ppb) gave a sharply upward parabolic response. Running a 1 ppb and 0.2 ppb check standard of my whole cal curve gave me a nice 1.x and 0.2x ppb result.
My mass library hit for TEL shows the 208, 237 and 266 peaks. I ran it as SIM in SIM/SCAN mode using 237 for quant and 208 for check. 1,4-dichlorobenzene-d4 is the internal standard. TEL comes out on a DB-VRX column at 16.17 minutes just before the last 4 peaks. Not a strong peak.
Instrument is Agilent 5973inert/6890N with an Edwards 230 turbopump and a fastscan side board and log amplifier. Running 20m x 0.18mm ID x 1um coating DB-VRX column. Purge gas is LN2; purge temperature 70C; Archon autosampler; Tekmar 3000 P&T; desorbing at 250C for 1 minute; trap is vocarb 3000. Valve body and transfer line 150C, inlet 200C. Source 150C Quad 230C. Slow T ramp from 40C to 230C; 1,2,3-trichlorobenzene is the last peak out at 16.80 minutes.
After a quicktune the next day my 40ppb is coming back as 22ppb for both my CCV and MS aliquots in a sample run.
If you have had good luck with TEL; please chime in.
I had just cleaned my source and was putting on an 8260 calibration. I got very good RSD's for my new calibration on the order of 5-7% RSD for all 79 regular compounds; except for TEL. It is distinctly non linear. I was able to fit a response factor curve (20, 40, 80, 120, 160 ppb) but any higher concentrations (240, 320, 400, 480 ppb) gave a sharply upward parabolic response. Running a 1 ppb and 0.2 ppb check standard of my whole cal curve gave me a nice 1.x and 0.2x ppb result.
My mass library hit for TEL shows the 208, 237 and 266 peaks. I ran it as SIM in SIM/SCAN mode using 237 for quant and 208 for check. 1,4-dichlorobenzene-d4 is the internal standard. TEL comes out on a DB-VRX column at 16.17 minutes just before the last 4 peaks. Not a strong peak.
Instrument is Agilent 5973inert/6890N with an Edwards 230 turbopump and a fastscan side board and log amplifier. Running 20m x 0.18mm ID x 1um coating DB-VRX column. Purge gas is LN2; purge temperature 70C; Archon autosampler; Tekmar 3000 P&T; desorbing at 250C for 1 minute; trap is vocarb 3000. Valve body and transfer line 150C, inlet 200C. Source 150C Quad 230C. Slow T ramp from 40C to 230C; 1,2,3-trichlorobenzene is the last peak out at 16.80 minutes.
After a quicktune the next day my 40ppb is coming back as 22ppb for both my CCV and MS aliquots in a sample run.
If you have had good luck with TEL; please chime in.