System: Agilent 1260 Infinity II
Detector: RID
Flow Rate: 0.9ml/min
Mobile Phase: DI Water (Milli-Q)
Pump: Quaternary pump (fed by mobile phase after Degasser, MCGV not connected - i.e. by-passed)
Sample loop: 100uL
Column compartment temperature: 50C
RID Temp: 35C
Column: LiChrospher 100 RP-8 (5um), LiChroCART 250-4, 4mm x 250mm
Injection volume: 20uL
Draw speed: 200 uL/min
Eject speed: 400uL/min
Delay between draw and eject: 1sec
I have recently been noticing a negative dip in my Chromatogram starting around 8.5 min and lasting through about 11in. The dip is only seen during an injection. It is not seen during a normal flow through of the mobile phase with the signal being monitored.
My mobile phase (Milli-Q DI Water) is degassed, but the sample solvent (which is also Milli-Q DI Water) is not.
The blanks that I have been injecting are vials filled with the Milli-Q DI Water.
I have attached an example chromatogram here: https://imgur.com/a/YFEbL2d
I have replaced multiple items within the System. Brief rundown: Sample Loop, Needle, Needle Seat, Rotor Seal, Purge Valve Filter, Pump Seals (2), Analytical Head Seal, manu-Cart NT Cartridge Holders, Phenomenex SecurityGuard and filter
All of these changes and no improvement. Based on the peak shape and dip, I am leaning towards gas/air being introduced with the injection but I do not know where. Any insight would be helpful.