Two possiblilities. 1. The sample has elements with large mass defect so monoisotopic mass is larger than nominal mass.
2. The mass spectrometer needs to be retuned to adjust the mass axis.
2 is most likely the case
I just thought of a third possibility. You didn't specify the type of instrument and ionization mode. If you are doing CI usually don't get molecular ion, you get molecular ion plus a proton.
2. The mass spectrometer needs to be retuned to adjust the mass axis.
2 is most likely the case
I just thought of a third possibility. You didn't specify the type of instrument and ionization mode. If you are doing CI usually don't get molecular ion, you get molecular ion plus a proton.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
You may be looking at something else. Give a bit more detail about what you are doign - instrument, compund, ionization thechnique.
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
If the instrument has passed a calibration (tuning), it can not be that big difference. You may try to re-tune the MS.
4 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 28 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 28 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science