Recently we are having problems in the lab with a mobile phase which is a mix of Phosphate buffer 10 mM with methanol (60:40).
The problem that we have is that sometimes when the mobile phase is purging, it starts as normally but then the pump does not have any more strength to purge the channel. What we do is discconnect the check valve and connect it again (without taking it out or changing it), and it suddenly has the strength to purge the channel.
Also, it has happend to us during a sequence. It is injecting normally and sudenly the pressure drops and it does not have any more strength to pump the mobile phase.
Before putting the mobile phase in the system, we make sure that the solution is mixed well and that it is cooled down.
This problem only happens when using this mobile phase, if another system is set up, this does not happen.
We do not consider that the mobile phase has a high salt concentration.
Does anyone know where this problem can come from? How could we solve it?
Thank you very much,