We have been quantifying MDMA with GC-FID since the beggining of the year; in our routine sequences we add some controls solutions, prepared with reference material standards.
Basic preventive maintenance (change of liner and septum) used to be monthly or once every six weeks, and there were no problems with the values of these controls (we obtained the expected results).
We analyse other drugs with the same chromatograph and we are not having any problem. However, "suddenly", one month ago, we have observed that a change of liner is necessary prior to our MDMA sequences if we want the expected values of our prepared MDMA standard solutions (even if we have made maintenance one week ago and the liner is apparently clean); otherwise, we will have lower values than the lower limit of acceptance.
The liner temp with this method is 230 ºC; with the other drugs is higher, and at the end of the sequences we run a standby method, where inlet temp goes to 150 ºC. We use split glass wool deactivated liners.
I dont'k know if these changes of temperature can affect the performance of the liner, but it was being our way of working until one month ago, without any problem, and now I'm questioning even the sense of living

If anyone has experience with these kind of analytes or liners, it would be very helpful for me if you have some idea.
Maybe it doesn't have nothing to do with inlet but with FID cleaning? (it is done annually and it will be done soon, it's been nearly 12 months since the last one). Maybe a cut of column is necessary? We should clean the syringe more often? (Injection volume 0,2 uL, syringe volume 10 uL).