Peak splitting
Posted: Wed Jul 17, 2024 8:40 pm
Recently I was troubleshooting peak splitting in UPLC H-Class using an Acquity® UPLC HSS T3 1.8 μm, 2.1 x 150 mm PDA e-lambda detector. MP A is a Buffer, MP B is 15% Buffer: 85% ACN. Gradient start at 97% A 3% B, Weak 90% H2O:ACN, Strong 50% ACN. I clean the instrument/column as Manufacturer recommend. After several trials troubleshooting the peak splitting with no results, instrument detector disconnects from Empower. I turned off the detector, then back on and now peak splitting it's not observed. For me makes no sense but I will like to know if someone experiment this and if it is something with detector reset.