Mobile phase: 10 mM phosphate, 700 mM NaCl, pH 6.1
Flow rate: 0.25 mL/min.
Problem description: the main peak experiences more tailing than before with specific column lots, in a concentration-dependent manner. We suspect that changes in resin lot might caused column lot-to-lot inconsistencies.
The interesting phenomenon is that when the antibody is diluted to 100 mg/mL (0.5 µL injection volume), the peak is more distorted than when the same is diluted to 25 mg/mL, from which 2 µL is injected (same mass load onto column).
When the sample is injected neat (~130 mg/mL), the peak tailing get even worse.
Dilution is performed with mobile phase.
Tailing causes overestimation of the LMW peak, eluting right after main peak.
Questions:
- How can one explain the concentration-dependent peak tailing, knowing the same mass load is injected?
- How can this be related to specific column lots?
- Can it be related to the extent of mobile phase with which the antibody is mixed, rather than the actual concentration of the antibody?
Instrument, methods, reagents, etc. were verified and are not in cause of this problem.
Thanks for your kind help!