A question about carryover.
We turn our GC off daily after testing - detector off, oven off, inlet off, leaving only makeup gas on.
Each morning on start up we run a water blank and we see 2 large peaks coming out at the end of the sequence as the temperature ramps. It is normal to see baseline disturbance but these are large well defined peaks.
However, we have observed that by leaving the inlet flow and inlet temperature on overnight (oven and detector still off) we don't see these peaks.
Also, these peaks only persist for the initial blank injection. They are gone by the second injection of the day.
Column flow is 4.58ml/min
Purge flow 5ml/min
Split flow 18ml/min
Column is ZB-624.
Septum and liner have been changed. This issue has persisted for around 2 months with the retention time of these ghost peaks very stable.
Any ideas for how and why this seems to be building up without flow would be appreciated.