Answering the question asked, it is likely N3+.
Answering the question that has not been asked, using N2 is one of the last things you want to do.
N3, that is one I never thought about, but it works out nicely, especially with so much nitrogen in the source. I doubt it was any of the others since it was coming through a cool column with only carrier flowing, and had just had H2 running through the system which should have knocked out most of any contamination present.
As for using N2, we are looking for alternatives to Helium since we currently use at least two tanks of Helium each week. As the price continues to increase alternatives are looking more and more attractive. Hydrogen being the preferred alternate carrier gas but working out the problems passing a BFB tune have not been easy. Getting the carrier flow down to 0.3ml/min gets it very close but will not pass tune consistently. I did see that someone (I believe it was from the Tel Aviv university) suggested trying a mixture of H2/N2 in which there is a ratio that would mimic the flow characteristics of He and be less reactive. I don't have the equipment to do a blend so haven't tried that yet.
In my little 3 hours or so experiment last week I found with N2 the chromatography works well except for the first freon peak which I expected, and the Ethyl Benzene, Xylenes peaks near the middle of the chromatogram which had shoulders on the tailing side almost to the point of being split peaks. I also had to raise the threshold since I was getting an across the scan range background of about 100-150 counts for almost all the masses which I did not expect along with about a 1-2% background level of m/z 42 when compared to m/z 69. This last problem only interfered with one compound so it was something I could work around if needed. Just above optimum linear velocity at 15cm/sec I was getting an analyzer pressure of about 6*10^5 torr, at 0.2ml/min using a 5973 with oil diffusion pump. Using a high performance turbo model would probably help with the sensitivity loss due to the higher pressure.
One plus was being able to begin the analysis run at a higher oven temperature and still have good retention of the early peaks, which allows the oven to reach the beginning temp faster on cool down.
I did notice that with N2 you get a larger sensitivity change with carrier flow change than with either H2 or He, most likely because of the size of the N2 molecule and changes in mean free path are greater for a smaller change in flow. Honestly if I had a spare instrument I could use just for experimentation I would dig out my old jet separator I have stored away and see how that would work with N2 or even N2/H2 carrier/makeup combinations.
I am switching from focusing on Hydrogen carrier for MS to simply Helium alternatives, whatever they may be.
Oh and thanks everyone for the suggestions on what my interference is and for keeping the discussion alive