post column split to set up Sulfur SimDis application D7807

Posted: Thu Jun 06, 2024 4:14 pm
by GGConsulting7
Hi, I know I could use a microfluidic splitter, but are there any other splitters I could (hopefully way less expensive) use to accomplish the split necessary for simultaneous Sulfur and SimDis detection? Thanks for any suggestions!

Re: post column split to set up Sulfur SimDis application D7807

Posted: Thu Jun 06, 2024 7:55 pm
by rb6banjo
I don't know the requirements of D7807 but if someone asked me about this type of application, I'd say this y-splitter might do the trick. Just adjust the length/diameter of the tubing from the Y to the respective detectors to get the split ratio you want.

https://www.restek.com/p/20277

Restek makes them in a few different dimensions to accommodate different column/tubing sizes.

I've used the press-fit ones with pretty good success. Just make sure the tubing cuts are flat. They usually seal up nicely.

Re: post column split to set up Sulfur SimDis application D7807

Posted: Tue Jun 11, 2024 1:50 pm
by GGConsulting7
thank you for the reply! Very kind of you!