HPLC mobile phase additive
Posted: Mon Jun 03, 2024 11:29 am
by Peter Apps
Many years ago I worked with HPLCs, and some methods included an additive to the mobile phase whose name I do not recall. It was a volatile organic with a strong and distinctive odour, almost certainly a sulfur compound, possibly added as an anti-oxidant or to stop analytes interacting with metal surfaces.
Does it ring a bell with anyone?
Re: HPLC mobile phase additive
Posted: Mon Jun 03, 2024 4:39 pm
by Multidimensional
Possibly TFA ? This was grossly overused in the early days of LC (and still is by many newbies).. Initially added at 0.1 to 1% to aqueous phases to prevent unwanted interaction with the lower quality silica supports used (Type A).
Posted: Mon Jun 03, 2024 4:59 pm
by Peter Apps
If TFA is trifluoroacetic acid, then no. This was a volatile.
Posted: Mon Jun 03, 2024 6:06 pm
by vmu
Triethylamine (TEA).
TFA is volatile too.
Posted: Mon Jun 03, 2024 7:59 pm
by Peter Apps
Not triethylamine - I am pretty sure that it was a sulfur compound, possibly a thiol of some kind?
Posted: Mon Jun 03, 2024 8:30 pm
by bcd_GCLCMSMS
Dithiothreitol (DTT)?
Used to prevent air oxidation of cysteine and the formation of disulfide bonds.
Posted: Tue Jun 04, 2024 4:17 am
by vmu
Thiodiglycol? However, it's BP is 165 C.
Posted: Tue Jun 04, 2024 9:58 am
by Peter Apps
Neither of those ring a loud bell - my recollection from more than 20 years ago is that the name was ........ ethanol.
Posted: Tue Jun 04, 2024 12:38 pm
by DR
not -ol, but maybe Heptafluorobutyric acid (HFBA)
or Methanesulphonic acid (MSA)?
Neither has the mercaptan nor bisulfate smell, but HFBA can clear a room...
Posted: Tue Jun 04, 2024 2:11 pm
by Hollow
one thing that come to mind would be mercaptoethanol.
Like used in the derivatization of amino acids with the OPA-method (o-phthalaldehyde derivatization)?
I wouldn't like to use it as a MP-additive but who knows?
Maybe it was a dedicated system with in-/online derivatization or other special method.
Posted: Tue Jun 04, 2024 5:15 pm
by Peter Apps
Mercapto ethanol does sound familiar, and OPA was being used for derivatizations. Quite possibly I misremember it being used as a mobile phase additive - I wasn;t actually doing the analyses.
The reason it comes up is that its odour is very similar to a potential predator chemical signal, whihc is how I now earn my daily crust.
Posted: Wed Jun 05, 2024 10:28 am
by lmh
Mercaptoethanol was used by molecular biologists all over the place, in protein gels, all reaction mixes, in anything. You could recognise any biological research institution from miles downwind by the general air of mercaptoethanol. Only the occasional genetics institute would manage to outsmell it, by a concentrated mix of mouse, and drosophila food. It was quite a warm, sulphurous smell that ended up pervading everything. I wouldn't be surprised if a biologist decided to put mercaptoethanol in their HPLC buffers; they probably offered a dilute solution of it to their girlfriends/boyfriends as perfume/aftershave back then. Dithiothreitol came as a great relief to many of us, as although it also has a distinctive smell, it's a lot less potent.
Posted: Wed Jun 05, 2024 11:57 am
by Peter Apps
Warm sulfurous is about right for the odour. This was in an analytical chemistry lab, for an analysis that was done only occasionally, ans so we never got ahbituated to the smell.
Posted: Wed Jun 05, 2024 2:38 pm
by DR
I left a job in part because I was starting to get habituated to a smell I really do not like.
Posted: Fri Jun 14, 2024 10:39 am
by wss
Would not surprise me if it was mercaptoethanol - we use it regularly as an additive in our LC system for mercury speciation using ICP-MS for detection. Distinctive warm sulfurous odour as you might expect...