PFAS Internal standard issue
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
We are experiencing high recovery of d3-N-EtFOSAA (internal standard), d5-N-EtFOSAA (surrogate) and N-EtFOSAA (analyte). Instrument calibrating fine and continuing calibration checks are fine. Only happening in extracted samples randomly. We initially thought it might be the preservative (Trizma) but are now leaning toward something to do the DI water or extraction cartridge. Has anyone experienced anything similar. thank you.
-
- Posts: 110
- Joined: Tue Oct 06, 2020 10:01 pm
Not sure what might be the issue, but I recall that the FOSAA compounds were the most sensitive to pH adjustment during extraction. If it was a bit too basic or acidic you could see a marked effect on the recovery. The exact pH range will depend on your method though.
How well is your calibration matched to your matrix (e.g. pH, solvent composition, etc.)?
How well is your calibration matched to your matrix (e.g. pH, solvent composition, etc.)?
-
- Posts: 3
- Joined: Wed Sep 22, 2010 3:58 pm
All samples and batch QC are preserved with Trizma to produce a pH near 7.0 in reagent water. It also removes free chlorine in finished drinking water which is what we are primarily analyzing. Since the calibration is not extracted it does not go through the extraction process.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
If it is a matrix effect you may need to run some preserved blank water through the extraction process then use that as the diluent for the calibration standards to make them matrix matched.
We have to do this for 525.3 since something in the matrix causes the responses for the last surrogate to almost double. If the standards are made in pure solvent then the extracted samples have almost 200% recovery. It is listed as a option in several drinking water methods, maybe it will help with this one.
We have to do this for 525.3 since something in the matrix causes the responses for the last surrogate to almost double. If the standards are made in pure solvent then the extracted samples have almost 200% recovery. It is listed as a option in several drinking water methods, maybe it will help with this one.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 25
- Joined: Mon Dec 11, 2023 4:02 pm
Hi all - sorry to open up an older thread but in recent travels, I've seen a couple of labs encounter this issue. The one common denominator has been the instrument (Agilent 6475). The most recent lab I chatted with even sent their failing extracts off to another lab to analyze on a different brand of MS/MS and those results came out fine (i.e the same vial of the same extract's Netfosaa's recovery comes out 150% at lab A and comes out 95% at lab B). This is even after all the troubleshooting & component testing I'd been able to help with on my end. It might be worth a service call, but I haven't seen any techs able to find the source of the issue. I wish I knew the solution - but I hope this offers some sort of insight.
-
- Posts: 110
- Joined: Tue Oct 06, 2020 10:01 pm
I've seen some very weird deviations in PFAS surrogate recoveries using Agilent 64xx triple quads... never entirely sure what causes it, and it seems to be intermittent and the period of stability does vary.
Current theory is that trace TFA from the acetonitrile and/or leaching out of the 1290 LC flow path components is causing some weird specific ion suppression in negative mode, charging effects, etc. which sometimes render the entire ion optics-quadrupole-detector path unusable without thorough cleaning. We run our systems very hard (24/7 uptime usually, with about 30% dedicated to flushing, maintenance, etc.)
Current theory is that trace TFA from the acetonitrile and/or leaching out of the 1290 LC flow path components is causing some weird specific ion suppression in negative mode, charging effects, etc. which sometimes render the entire ion optics-quadrupole-detector path unusable without thorough cleaning. We run our systems very hard (24/7 uptime usually, with about 30% dedicated to flushing, maintenance, etc.)
6 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science