LCMS Inconsistent Signal Ratios compared to ISS
Posted: Thu May 30, 2024 5:04 pm
I have something of an issue with our LCMS system which I haven't been able to find any explanation for. I'm attempting to run amphetamine and methamphetamine in a Waters system, using a 50x2.7 column with 2.1 uM beads coated with C18, then to a Xevo-TQs tandem mass spectrometer in MSMS mode with the ESI+ spray. With the amphetamine and methamphetamine, I'm using amphetamine-d5 and methamphetamine-d5 respectively as internal standards. I have water 0.1% Formic Acid and 5 mM ammonium formate as solution A, and Acetonitrile @ 0.1% Formic acid as solution B. The solvent starts at 90:10 initial and ramps up to 0:100 over the course of about 10 minutes.
I'll have a sample of 90:10 Water:ACN spiked with my compounds and their associated ISS, and I'll run that sample multiple times. When I do, I see the signals of the amphetamine and methamphetamine generally decrease with each run. That wouldn't be such a big problem, except that their ISS signals don't decrease proportionally. The ratio of my amphetamine signal to it's ISS is generally decreasing by about 30% (even after letting my column condition at the initial settings for about an hour or so). Since this is the same sample run multiple times, that ratio should be roughly constant, but it's varying dramatically, and non-randomly. The ratio is going down with each run.
I've tried again running the Xevo in just MS mode (no fragmentation), and then the problem seems to go away, which kind of baffles me. I am completely at a loss to explain why this is happening am hoping to get some clues for additional things to look at with my system/protocol, etc.
I'll have a sample of 90:10 Water:ACN spiked with my compounds and their associated ISS, and I'll run that sample multiple times. When I do, I see the signals of the amphetamine and methamphetamine generally decrease with each run. That wouldn't be such a big problem, except that their ISS signals don't decrease proportionally. The ratio of my amphetamine signal to it's ISS is generally decreasing by about 30% (even after letting my column condition at the initial settings for about an hour or so). Since this is the same sample run multiple times, that ratio should be roughly constant, but it's varying dramatically, and non-randomly. The ratio is going down with each run.
I've tried again running the Xevo in just MS mode (no fragmentation), and then the problem seems to go away, which kind of baffles me. I am completely at a loss to explain why this is happening am hoping to get some clues for additional things to look at with my system/protocol, etc.