PFAS-free manual SPE vacuum manifold
Posted: Wed May 29, 2024 1:14 pm
Does anyone know of a confirmed PFAS contamination-free brand/model manual vacuum manifolds that are in use for EPA 533, EPA 537.1 and EPA 1633?
We have tested a couple (including the 24-port version of the EPA 533 recommend model) and they all contain significant background contamination of at least 1 or 2 analytes found in those methods. Possibly some of the parts have changed since method was written. I know most manifolds contain PTFE parts, but the extracts only contact the Luer ports, valves and guide tubing to the polypropylene collection tubes.
Our production automated systems are actually far cleaner than the manual SPE devices we have tested.
We have tested a couple (including the 24-port version of the EPA 533 recommend model) and they all contain significant background contamination of at least 1 or 2 analytes found in those methods. Possibly some of the parts have changed since method was written. I know most manifolds contain PTFE parts, but the extracts only contact the Luer ports, valves and guide tubing to the polypropylene collection tubes.