- Posts: 83
- Joined: Wed Mar 25, 2020 8:52 am
I work with a waters tripple quadrupole ms instrument and was in the process of verifying the ion transmission using waters own setup solution. I have done full scan on MS1 and a full scan using MS2 (no fragmentation collision gas off). I hade about twice as high peak height on MS1 for scan vs using the MS2 to scan. All parameters the same only difference is what quad is doing the scanning and the other just doing ion transmission.
Both scans look fine in terms of signal, calibration and resolution. The MS1 scan did give better signal by around 2 times the signal using MS2 for scanning. That was about the only difference I could see. They quads have a bit different resolution so some difference in signal intensity is to be expected as it goes up when resolution goes down, but they were fairly similar in resolution.
1) Is this an appropriate test for evaluating MS1 and MS2 ion transmission?
2) Is there any reason why MS1 is the preferred quadrupole for making the scan and MS2 as passive transmission of ions, and if so why (or was my result with MS1 being better just a coincidence)?
I guess I should know the answer to this but since i don't I would appreciate if someone could enlighten me.
Thanks!