Waters 2695 HPLC Degasser Pressure Fault

Posted: Fri Apr 12, 2024 8:28 pm
by zpowers
Hi All,

My Waters 2695 HPLC is giving me a desgasser pressure fault (4.7 psi) and the liquid coming out of the system is full of air. I've primed the pumps, but the problem persists. Any ideas?

Zack

Re: Waters 2695 HPLC Degasser Pressure Fault

Posted: Sun Apr 14, 2024 7:37 pm
by Multidimensional
Your integrated Alliance vacuum degasser has serious damage requiring replacement or professional service. If well maintained, the degasser modules are designed to last a maximum of five years (people often ignore these modules resulting in internal flow path contamination). *Liquid may be visible exiting the exhaust line, red in color powder observed by the vacuum pump or perhaps a knocking (or "howl") may be heard coming from the pump (all signs of serious damage). Here is a link to a free article with troubleshooting information: "WATERS ALLIANCE & ACQUITY UPLC DEGASSER ERRORS : https://www.hplctools.com/waters_degass ... ooting.htm