- Posts: 19
- Joined: Thu Nov 16, 2023 7:25 pm
We are running an Agilent 1100 series HPLC with an RI detector. My historical experience with HPLCs was with a system that ran 24/7 in a corn syrup refinery that used water as the mobile phase and so we really never had to equilibrate the systems except for maintenance. It has also been so long ago that I can't remember how long we would leave the system alone to equilibrate. On our current system, we are running a 75% acetonitrile mobile phase and so we shut the system off/turn it on whenever we are using it. This is a new for us/used system and we have not worked on it enough to know what to expect for equilibration time. I ran my calibration curve the other day and it was less than stellar as far as baseline stability/reproducing similar results between injections. I am used to being able to reproduce results within ~0.15% all day every day with the Waters system I used at my previous job that ran 24/7 365 days a week but this one has not been able to give me anything I would consider consistent within ~1.5%. Any other tips or things to look at would be helpful
Pump is running 1.4 mL/min
Column @ 35C
Detector @ 40C
Column is a Waters WAT084038 silica amino reverse phase 3.9mm(id)/300mm(length)