Agilent G1888 Headspace.
Posted: Sat Mar 02, 2024 1:14 am
We were attempting to replicate a paper analyzing hand sanitizer for residual solvents/adulterants and observed something I believe that could be an issue of the headspace autosampler. I was having my students use cyclohexane for recovery calculations. They spiked their samples so that the concentration would double, but it did not. Some groups saw about a 1.2 increase in response and saw a decrease in response.
One thing that I did notice, is that the paper didn't specify a vial pressurization or carrier gas pressure. Right now our system is "set" to 15 and 25 psi, respectively. I put set in quote marks because that's what the headspace method is. However, when I checked the actual pressures on the headspace, I observed them to be 14.8 and 17.3 (vial pressurization and carrier).
Our GC uses EPC to control vial pressurization and MPC for carrier pressure so that we can disconnect the transfer line from the back inlet. The GC uses an EPC to supply the carrier pressure and is set to 65 in the method/instrument. The He cylinder is set to an output of ~80 psi.
Ignoring that they're students and prepared everything correctly (ha), this has me concerned. I've come up with the following plan:
Do manual injections to see if the area doubles.
Check for leaks on the HS.
Repeat the analysis using a different spike.
Use methanol, acetone and 1-propanol to assess injector linearity.
Is there anything else I can do to assess the headspace?
One thing that I did notice, is that the paper didn't specify a vial pressurization or carrier gas pressure. Right now our system is "set" to 15 and 25 psi, respectively. I put set in quote marks because that's what the headspace method is. However, when I checked the actual pressures on the headspace, I observed them to be 14.8 and 17.3 (vial pressurization and carrier).
Our GC uses EPC to control vial pressurization and MPC for carrier pressure so that we can disconnect the transfer line from the back inlet. The GC uses an EPC to supply the carrier pressure and is set to 65 in the method/instrument. The He cylinder is set to an output of ~80 psi.
Ignoring that they're students and prepared everything correctly (ha), this has me concerned. I've come up with the following plan:
Do manual injections to see if the area doubles.
Check for leaks on the HS.
Repeat the analysis using a different spike.
Use methanol, acetone and 1-propanol to assess injector linearity.
Is there anything else I can do to assess the headspace?