Limit of detection and error in analysis
Posted: Wed Feb 03, 2010 8:52 pm
Hey Guys,
I've searched the site to get an idea of how to correctly determine the detection limits and introduce error bars on my data. Unfortunately i've encountered a few problems.
For example
This guys states that:
"from the linearity curve find out residual standard deviation and slope.
Now one can derive
LOD = 3.3*residual stdandard deviation / slope.
LOQ = 10*residual stdandard deviation / slope.
FOR finding out resdiual standard deviation one can use MS-EXCEL application. for the selected two sets of variables follow =styex(Y1:Yn, X1:Xn)."
To calibrate the GC i've run two standards. 1) at 1% methane/ He balance and a 15 ppm methane/ He balance standard. Since i plot the concentration vs peak area, my slope is around 0.0003x. I ofcourse cannot get the residual standard deviation as well since i have only two calibration standards. Does anyone have a good technique to figure out error bars in this type of situation. I think for LOD i will take the 3 times the noise level and for LOQ i will take 10 times the noise level.
Thanks
For example
Now one can derive
Thanks