Waters 486 Failure Code 401

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I have had this detector for about 6 years (purchased used). I had to change the lamp about a year later, and it has run perfectly since. This morning, I had to power it off to change out a UPS; and when I turned it back on, it started clicking when it went into calibration step and ended with a Fail 401 error. I have a feeling that the problem is with the monochromator drive. Fortunately, I have another working 486 to substitute, although the sample and reference energy is low. I'm not without an instrument. Is this an easy fix?

Thank you,

John
As I remember , there is a button lithium cell on the electronic board , which stores the calibration value.

You may replace it , and make the calibration again.
Fill the flow cell with fresh acetonitrile , before calibration.
2 posts Page 1 of 1

Who is online

In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry