Can a defective precolumn fitting damage UHPLC tube fitting?
Posted: Fri Jan 26, 2024 1:39 pm
I noticed a higher than normal back pressure for a couple of batches on my UHPLC, but the chromatography looked good. The somewhat higher pressure was present since the previous pre-column install, I assumed that the pre-column was getting occluded with fine particulate from the last few batches, so I swapped it for a new one from the same box (Box of 3). (Note: these are true pre-columns with their own, new fittings, not a replaceable filter in a holder.)
I reassembled the unit and started it. The pressure soared up instantly and the pump stopped at the high pressure limit.
Not sure what's wrong, so I removed the pre-column and column as a unit and tried to pump solvent through the bare fitting to check the pressure, but the tubing appears to be completely blocked at the end fitting from the UHPLC tubing. No flow at all! Visually, I can see nothing wrong with the threads or the ferrule, but I haven't used a magnifier or microscope to look at the end.
Is it possible that I got a bad lot of pre-columns and they somehow damaged the tubing fitting to the column?
Any repair possible or order a new section of SS tubing?