Q. 1) spike is a "bad" word in MS circles. It implies a noise event/electrical discharge etc. Here you are expecting 4 responses, if you set up 4 MRM expts in the datafile. They should occur chromatographically aligned beneath each other.
Q. 2) I cannot tell you exactly what the scan acquisition details in the HELP files were. However, I can point out some details on their printouts that you should be aware of,
a) SM 5G means that a 5-point gaussian smooth has been applied to the mass chromatogram. This done to make the peak look "GOOD" with a continuous peak shape. As a matter of interest, try applying in succession a 1-point, 3-point, 5-point Gaussian smooth to your own data file; note how the very spiky chromatogram becomes smoother and smoother, and how the area of the peak (peak counts) changes. Try the effect of other smoothing algorithms and compare.
b) The analyst is looking at the loss of 45 amu from the ion at m/z 465.3.
465.3 - 45 = 420.3; a peak centroided at m/z 420.3 would be 419.8 to 420.8 (1 amu wide at unit mass resolution).
However, the analyst has set his detection range at m/z 419.3 to 421.3 (this is 2-amu wide ) and gives a greater response than the 1-amu window above.
c) The same wider than 1-amu detection window is used for the D4 internal std. shown in the lower HELP file.
Similarly, the analyst can widen the width of the parent ion that undergoes MS/MS fragmentation; instead of m/z 464.8 to 465.8 (as above) for the 12C, 1H,14N etc molecule [M+H] , the 2H, 15N, 15N isotopic molecule at [M+H+1] could be included to increase the response.
Data smoothing and using expanded width detection windows are widely used, esp. in low-level quantitation by MS.
Note also that the deutero- analog mass chromatogram maximises just 1-2 scans ahead of the compound being quantified---this is a general effect.
Looking at the mass chromatogram profiles in the HELP files it looks like most of the signal lies in the time range 4.8 to 5.0 min, i.e. 12 secs wide. you need a certain minimum number of scans to define the chromatographic peak shape.
1-secs scan times would give you only about 3 scans at each of 4 x MRM expts. NOT ENOUGH
0.5 sec.......about 6 scans NOT ENOUGH
0.25 sec.....about 12 scans JUST enough ?/!
0.1 sec..... about 30 scans EXCELLENT to define peak shape.
Suggest that to get a feel for this that you run 3 expts successively with your 4 x MRM at each of the time settings.
Find the best time setting for data acquisition, then run 3 expts successively with the parent ion set at m/z 609.5 (1-amu wide), then 2-amu wide then 3-amu wide. See how the intensity of m/z 174 changes.
NB don't forget to expand m/z 174 if you widen the mass width on the parent ion.
d) Looking at your datafile it looks like you have set an extremely low width (174.001 - 173.999 = 0.002 amu). If the daughter ion actually is at m/z 174.0, you should set m/z 173.5 to 174.5 scan range. if at m/z 174.2, then m/z 173.7 to 174.7
NB There will be cases when, because of interfering compounds in the sample solution you may have to use very narrow mass windows.
We look forward to seeing your progress.
James Ball has a LOT of expertise in this area and if he posts, you should listen!
Regards,
JMB
THANK YOU SOOOOO MUCH for this!!
I'll be sure to avoid the term "spike." It was a bad word when doing Raman in grad school too, haha.
Thank you also for the breakdown of the scan times and number of scans. I've seen anywhere between 12 - 20 scans/peak is recommended. Thermo said 14 when I gave them a call regarding this. I will definitely give the experiments a try at the different scan times to get a feel for it!
The 0.002 amu width seemed crazy small to me as well. That was the default value in the HELP file. Your suggestion of 0.5 amu makes entirely more sense to me than such a narrow window - at least for now while trying to get a feel for this.