Page 1 of 1
Rounded rules and complies
Posted: Mon Jan 15, 2024 12:52 pm
by crimelist
Hi...
I received a complaint. The first of this type after many years of work.
Request:
Specification limit: ≤ 0.4 %
Result 0.42%
Is it compliant in your opinion? Or not?
Re: Rounded rules and complies
Posted: Mon Jan 15, 2024 8:26 pm
by DR
if spec were written 0.40, there's an issue. Your result is in spec (barely), tell the complainer to rewrite the spec w/ justification if they want 0.42 to be OOS.
Posted: Tue Jan 16, 2024 1:36 pm
by Consumer Products Guy
I agree with DR. Within specifications.
Had it been written as Specification limit: ≤ 0.40 %, then would be OOS.
Posted: Wed Jan 17, 2024 3:10 pm
by bunnahabhain
I also agree with DR. To avoid this type of complaint, fix your report format to round the required number of digits...
Posted: Fri Oct 25, 2024 3:18 pm
by smalldots
for the spec of NMT 0.4%, you can go up to 0.44% technically.