difficulty with method for detecting AAT
Posted: Tue Jan 09, 2024 12:13 pm
Hi
I am trying to follow this method https://www.agilent.com/cs/library/appl ... 5169EN.pdf
I am only doing the HPLC part and was hoping that this would give the separation needed, unfortunately I am not having much luck
HPLC is set up as method states except for a few changes

https://ibb.co/t4ghwWj
- column: Poroshell 120 EC-C18 4.6 x 100mm 2.7 Micron (this is one we already have onsite and thought it would be similar enough but clearly i was wrong)
- mobile phase A 0.1% formic acid in water
- std prep: 100mg/l AAT std in Acetonitrile
- bx prep: red sample spiked with 100mg/l AAT std in acetonitrile
1st run
std:
https://ibb.co/j6YMLJm
bx:
https://ibb.co/jzZYR7w
std gave either a 2nd peak immediately after or 1 badly shaped peak with an unexplained drop to zero mid peak
bx gave a peak that dropped below zero, not sure why this may be as the internet said this is usually caused by a mismatch of your mobile phase and dilutant solvent, but both are acetonitrile.
I then ran repeats of my std to see if the peak shapes were consistent, which the shape and retention time were but the peak heights varied between injections, no pattern ie increasing peak height suggesting sample being retained.

https://ibb.co/bbcg4Sz
I then made another dilution of the std at 5mg/l and changed the injection volume to 10μl in case the hplc was struggling with the 0.5μl injection volume, and again ran 4 repeats but this has just resulted in more confusion. only the 2nd injection looks like the previous ones and the remaining 3 still go negative but have a peak later in the sequence not present in previous runs.

https://ibb.co/yBvjdxQ
please could I get some advice of what i need to change to get the method to work or is it just a fruitless endeavour and i should stop wasting my time. or if any of you have a working method for detecting low levels (<0.1%) of AAT (o-aminoazotoluene) from liquid fuel dyes youd be willing the share.
Thank you
I am trying to follow this method https://www.agilent.com/cs/library/appl ... 5169EN.pdf
I am only doing the HPLC part and was hoping that this would give the separation needed, unfortunately I am not having much luck
HPLC is set up as method states except for a few changes
https://ibb.co/t4ghwWj
- column: Poroshell 120 EC-C18 4.6 x 100mm 2.7 Micron (this is one we already have onsite and thought it would be similar enough but clearly i was wrong)
- mobile phase A 0.1% formic acid in water
- std prep: 100mg/l AAT std in Acetonitrile
- bx prep: red sample spiked with 100mg/l AAT std in acetonitrile
1st run
std:
https://ibb.co/j6YMLJm
bx:
https://ibb.co/jzZYR7w
std gave either a 2nd peak immediately after or 1 badly shaped peak with an unexplained drop to zero mid peak
bx gave a peak that dropped below zero, not sure why this may be as the internet said this is usually caused by a mismatch of your mobile phase and dilutant solvent, but both are acetonitrile.
I then ran repeats of my std to see if the peak shapes were consistent, which the shape and retention time were but the peak heights varied between injections, no pattern ie increasing peak height suggesting sample being retained.
https://ibb.co/bbcg4Sz
I then made another dilution of the std at 5mg/l and changed the injection volume to 10μl in case the hplc was struggling with the 0.5μl injection volume, and again ran 4 repeats but this has just resulted in more confusion. only the 2nd injection looks like the previous ones and the remaining 3 still go negative but have a peak later in the sequence not present in previous runs.
https://ibb.co/yBvjdxQ
please could I get some advice of what i need to change to get the method to work or is it just a fruitless endeavour and i should stop wasting my time. or if any of you have a working method for detecting low levels (<0.1%) of AAT (o-aminoazotoluene) from liquid fuel dyes youd be willing the share.
Thank you