Shimadzu GCMS autotune failing 502 intensity

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I have a GCMS=QP2020NX from shimadzu that started failing a 502 check after changing the column. I cleaned the source, changed the filaments, cleaned the lens stack, cleaned the rods. Still nothing.

Any ideas?
I have a GCMS=QP2020NX from shimadzu that started failing a 502 check after changing the column. I cleaned the source, changed the filaments, cleaned the lens stack, cleaned the rods. Still nothing.

Any ideas?
Revert older column
Check the position of the column at the ion source. It is sensitive to the geometry.
3 posts Page 1 of 1