Errors occured during GCMS Analysis

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Can anyone help me resolve these issues?
I have encountered the following errors during GCMS analysis:
1) CAR1 AFC leaks.
2) CAR1 primary pressure is out of range
3) Carrier gas does not flow sufficiently
4) Vacuum Auto Startup sequence has stopped.
Need to know which model and vendor the GCMS is to know what the errors mean. Each will have its own error codes.
The past is there to guide us into the future, not to dwell in.
Can anyone help me resolve these issues?
I have encountered the following errors during GCMS analysis:
1) CAR1 AFC leaks.
2) CAR1 primary pressure is out of range
3) Carrier gas does not flow sufficiently
4) Vacuum Auto Startup sequence has stopped.
You have the answer already - number 1) - there is a leak. The rest is a consequence of that.

Fix it.
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry