HS/GC stops collecting data on vial 8 and i dont know why
Posted: Tue Feb 02, 2010 3:48 pm
I have this strange problem. Let me first state that I am using a G1888A headspace sampler with an Agilent 6890 GC.
So, I ran a sample set that had 13 vials twice. Both times the first seven vials are completed and the eighth vial showed 'missing data' in empower. Also, the sample set was:
vial 1 Air
vial 2 blank
vial 3-7 standard
vial 8 blank
vial 9 sample endogenous
vial 10 spiked sample
vial 11 blank
vial 12 standard
vial 13 blank
For vial 6 & 7 the chromatogram does not show any standard peaks.
The run time is 20min but the vial equilibration time in headspace is 40min.
GC oven: 40C (2min) ---(10C/min)---->220 (0min)
This exact situation occurred twice.
So far what I've tried to remedy the problem is changing the GC oven program to
40C (2min) ---(10C/min)---->220 (2min) --------(20C/min)----->40C
to see if the problem was that the GC was not ready when the headspace injected.
I'll post if this worked later. Any help would be appreciated.
So, I ran a sample set that had 13 vials twice. Both times the first seven vials are completed and the eighth vial showed 'missing data' in empower. Also, the sample set was:
vial 1 Air
vial 2 blank
vial 3-7 standard
vial 8 blank
vial 9 sample endogenous
vial 10 spiked sample
vial 11 blank
vial 12 standard
vial 13 blank
For vial 6 & 7 the chromatogram does not show any standard peaks.
The run time is 20min but the vial equilibration time in headspace is 40min.
GC oven: 40C (2min) ---(10C/min)---->220 (0min)
This exact situation occurred twice.
So far what I've tried to remedy the problem is changing the GC oven program to
40C (2min) ---(10C/min)---->220 (2min) --------(20C/min)----->40C
to see if the problem was that the GC was not ready when the headspace injected.
I'll post if this worked later. Any help would be appreciated.