How tu purge air line in Agilent 6850 GC

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello!

Our compressed air tank has run out but there is still 40 psi of residual pressure in the regulator. I don't think it's safe to try and change the regulator with 40 psi in it. Our only outlet is going straight to the 6850 GC. Is there a way to use the GC6850 software to purge the residual pressure, making it safe to change the regulator?

Thanks!
You may simply loosen a nut connecting the air tubing to the GC or the regulator , until the pressure is released.
Are you talking about removing the regulator from the cylinder with 40psi on the high side?

We normally change cylinders when the pressure drops below 500psi, and we just close the valve and remove the regulator. There is a quick hiss as the pressure is released but it does no damage to the regulator and is safe for the operator. We even do this with Hydrogen cylinders with no problems, just be sure the cylinder valve is closed first.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry