Corona CAD carryover qualification
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
We be normally qualifying the HP1200 carryover using the UV detector. Howevers, the UV detector be out of available so we look at qualify carryover using CAD which also attached to HP1200. However, response from Caffine is very low. Is there anyone knowing more suitable alternative moecule to assess carryover on ChAD?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
The evaporative detectors, as a group, do not show nearly the variation in response of a UV detector. In essence, they detect dust; dust is dust, and you get about the same response regardless of the composition.
My understanding is that the detection limit of the CAD is down in the low nanogram range. I you are doing significantly worse than that, you might try tweaking the parameters (nebulizer gas flow and temperature, mobile phase flow, and perhaps mobile phase viscosity). If you have already done that, your only option is to increase the mass injected.
My understanding is that the detection limit of the CAD is down in the low nanogram range. I you are doing significantly worse than that, you might try tweaking the parameters (nebulizer gas flow and temperature, mobile phase flow, and perhaps mobile phase viscosity). If you have already done that, your only option is to increase the mass injected.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Is this now accepted GLP, or whatever, procedure to check the carryover with a substance other than the analytes? In my experience there has been a strong dependence of carryover on type of analyte.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
In OQ context, it is necessary and accepted to use some kind of generic standard/test material. Could be caffeine, uracil or whatever makes sense.
In PQ context, however, the carryover is/must be tested using the actual analyte.
Best Regards
In PQ context, however, the carryover is/must be tested using the actual analyte.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
4 posts
Page 1 of 1
Who is online
In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 13 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 13 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science