Tailing solvent peak - in split mode
Posted: Sat Jan 30, 2010 2:23 am
Hi all,
We are running a C14 & C15 analysis (roughly ppm levels) to demonstrate the use of internal standards in GC-MS analysis. The solvent is hexane and we are using a 20:1 split at 180C.
We have noticed recently that the hexane solvent tails badly into the two analytes peaks which come off around 2 and 3 minutes on our 30 m HP5MS column. The liner is a split liner and the injection port temperature is 250C. This method used to work with no problems. The chromatography of the C14 and C15 peaks is normal other than they lie on the hexane peak of course.
Today whilst running another split analysis of C7-C12 hydrocarbons in hexane with the same column and liner I noticed that the hexane was tailing for over 5 minutes. It was possible to identify hexane as the major contributor to the baseline even when the higher hydrocarbons such as C10, C11 and C12 were being eluted. Again this was never previously the case. Again the peak shapes of the analyte peaks are okay
Any suggestions as what to try. My first instinct is obviously to change the liner but the current liner was clean on inspection perhaps indicating something else. Am I on the right track if I think that the gold seal and/or the split vent may be the next areas to look at.
Thanks in advance
Kevin
We are running a C14 & C15 analysis (roughly ppm levels) to demonstrate the use of internal standards in GC-MS analysis. The solvent is hexane and we are using a 20:1 split at 180C.
We have noticed recently that the hexane solvent tails badly into the two analytes peaks which come off around 2 and 3 minutes on our 30 m HP5MS column. The liner is a split liner and the injection port temperature is 250C. This method used to work with no problems. The chromatography of the C14 and C15 peaks is normal other than they lie on the hexane peak of course.
Today whilst running another split analysis of C7-C12 hydrocarbons in hexane with the same column and liner I noticed that the hexane was tailing for over 5 minutes. It was possible to identify hexane as the major contributor to the baseline even when the higher hydrocarbons such as C10, C11 and C12 were being eluted. Again this was never previously the case. Again the peak shapes of the analyte peaks are okay
Any suggestions as what to try. My first instinct is obviously to change the liner but the current liner was clean on inspection perhaps indicating something else. Am I on the right track if I think that the gold seal and/or the split vent may be the next areas to look at.
Thanks in advance
Kevin