New column woes

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
I installed a new column on our Agilent triple quad today and realized after about 30 seconds that it was connected to the detector. I switched over to waste, but I now have a background signal I cannot get rid of. I have been searching to find out what might have gotten into the MS and what to do about it, but can only find documents saying "When installing a new column, do NOT connect directly to detector." Can anyone tell me what I might have accidentally done to the MS and how I might be able to fix it? Your help would be greatly appreciated.

what's the column?

It is a Zorbax Eclipse Plus C18 RRHT 2.1x10mm 1.8 micron column.

The instruction makes no sense in the context of this column. If you think that the problem is related to the column, call Agilent.

http://www.chem.agilent.com/en-US/Produ ... 66609.aspx

It doesn't really make sense that Agilent would make that statement, but they sure do. I've only heard the 'don't connect the column' caveat when reversing a column to attempt cleaning or clearing a blocked inlet frit. Perhaps Agilent is worried about fines clearing the column end frit and accumulating in the detector, hence the warning?

I am under the understanding that these columns (like most HPLC columns) are pre-run and even an individualized chromatography report supplied with the column. So it really shouldn't be any chemical or particle contamination, should it, could you be that unlucky? How long have you been trying to flush the detector, and with what?
6 posts Page 1 of 1

Who is online

In total there are 39 users online :: 2 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google Adsense [Bot], Semrush [Bot] and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry