Page 1 of 2
one question about my chromtogram
Posted: Sat Jan 23, 2010 9:49 am
by jiang295
this is extract from bread. why baseline has so huge bump? what are underneath in baseline. I check control, baseline is very flat. so there wont be any problem associate with detector or column. any idea?
Posted: Sat Jan 23, 2010 2:47 pm
by bisnettrj2
Can you supply your instrument conditions? Mobile phase constituents? Injection volume? Sample diluent?
Posted: Sat Jan 23, 2010 4:57 pm
by jiang295
column 250mm*21mm C18.
mobile phase methanol and 0,1% formic acid.
injection volumn: 2ml.
injection mass: 100mg
elution condition: 5%-90% 30min and hold 90% 8min
Posted: Sat Jan 23, 2010 7:51 pm
by bisnettrj2
Is the mass of sample injected equivalent to the standard that you injected that did not show the baseline hump? Was this sample injected right after the standard? How long do you re-equilibrate your column after the 8 minute hold at 90% MeOH? And what is your sample diluted in? 95:5 0.1%formic acid:methanol?
Posted: Sat Jan 23, 2010 8:20 pm
by Hollow
I wouldn't search too far and blame something to be wrong with the system.
As you state this is an extract from bread, I would expect it to be some polymeric compounds from the sample itself.
I've seen something similiar with some plant extracts rich on polyphenolic compounds.
What was you sample prep?
Posted: Sat Jan 23, 2010 9:13 pm
by jiang295
I went throught SPE, and using 10% methanol and 100% methanol to stepwise elute.
only collect fraction elute with 100% methanol. so polysaccharide and protein will be excluded from my sample.
I did not consider polyphenolic compounds, that is quite possible.
Posted: Sat Jan 23, 2010 9:17 pm
by jiang295
did you prove that those are polyphenolic compounds from plant extract?
Posted: Sat Jan 23, 2010 9:44 pm
by bisnettrj2
Can you post a chromatogram of a standard that equates to a similar concentration as your sample extract, so we can see the difference between your 'good' chromatogram and the chromatogram that you have posted?
Posted: Sat Jan 23, 2010 11:34 pm
by jiang295
why do you think so? you think that is because of mass overload?
Posted: Sat Jan 23, 2010 11:43 pm
by bisnettrj2
Not necessarily. I just think it will be helpful to see what you consider to be a good chromatogram in this analysis, so we have a means of comparison. I suspect that Hollow is right, that it is something in your sample that isn't present in your standards, but I think it will help to rule out any instrument or inherent analysis issues.
Also, you say your SPE preparation consists of eluting the portion of interest from the SPE with 100% methanol. Do you inject 100% methanol, or do you dilute your sample with 0.1% formic acid before injection?
Posted: Sun Jan 24, 2010 1:41 am
by jiang295
I really appreciate your help. thanks. i just want to know reason why you want to do that.
i get rid of all methanol and redissolve in water before injection.
Posted: Sun Jan 24, 2010 3:19 pm
by Hollow
proved? more or less
just considered some coindices.
- injection of a "pure" tannin sample showed also a bump like this
- the plant is known to contain such substances
- by hydrolysis with conc HCl the bump decreased
- with HCl the solutions often turns dark red -> (pro)anthocyanes
- incubating the sample with skin powder or bentonite reduces also this bump (unfortunately also other peaks)
for your bread extract:
What about Lignines and/or starch?
Posted: Sun Jan 24, 2010 6:22 pm
by Uwe Neue
There may be ways to suppress the elution of polymers from SPE by using selective solvents for elution. If you do not know what this is, this may amount to a trial-and-error exercise. I can help you with some ideas using Oasis SPE packings.
Posted: Sun Jan 24, 2010 8:28 pm
by jiang295
i guess it is Lignines.
does starch retain in the C18 since they are so hydrophilic?
Posted: Sun Jan 24, 2010 8:40 pm
by jiang295
sure, so there is unique characteristic of this SPE material.