PREP Column with peak splitting problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hi,
i'm using an AGILENT Prep HT Eclipse XDB-c18 (150x21.2-5µm) column for purification of several kinds of molécules.
The problem is that i got peak splitting with a ratio of 20-80 for the amount of split with every kind of product, so the sample solvent can't be the cause.
I would like to know if you can tell me a recovery procedure to fix the problem or maybe the column is "dead" :-[ ....
I use a column guard of the same stationnary phase but it seems it has not protected the column ....
I work for an R&D outsourcing laboratory in drugs research so i can't tell you what was the product i use when the problem apeared...., the eluent system is ACN-water or sometime TFA 0.1% aq. buffer.
Thanks

what are the suggestions from Agilent technical persons?

Column stability on prep columns is a non-trivial business. I suggest to look at another vendor that has solved the problem, i.e. Waters.

I am usually not that bland, but there is really a large difference in the technology between different vendors, and it is worth your while to investigate, especially considering the price of such columns.

The problem seems to be solved:
I used the column inverted, washed with Water and ACN 95-5 to 5-95 and 5 injections of 200µL of DMSO.... so for the moment, i don't have any peak splitting, just a little tailing....
Thanks for your replys.

Nicolas

Nicolas,
I’m also having the peak splitting problem. The column that I’m using is Bondapak phenyl, 3.9 x 300 mm 10 . Why did you inject DMSO five times after washing your column?
Thanks,
LV

Provetech,
I might suggest a possible reason for the peak splitting might be partial clogging of the inlet frit on the column. I have seen this split peaks before, and lloks very much like a column void. Maybe your pressure increased slightly due to some insoluables or ??? Anyway just another thought about the splitting.

Regards,
Mark
Mark
6 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry