Peak shape change vs time
Posted: Mon Jan 18, 2010 6:37 pm
Fellow chromatographers,
i've troubles optimising our analytical method : the peaks of all the target compounds are broadening (peak fronting) in function of time (see the picture below).

We have an Agilent 1100 LC/MS (MWD + Ion Trap) and as you can see, i use a gradient system with following settings @ a flow rate of 0.4 ml/min:
00.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
08.00min 60% H2O / 40% Acetonitrile + 0.1% formic acid
08.50min 0% H2O / 100% Acetonitrile + 0.1% formic acid
09.50min 0% H2O / 100% Acetonitrile + 0.1% formic acid
10.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
15.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
The column is a 2.1 mm x 20 mm C18, temperature controlled @ 35.0 C and with an injection volume of 3 µl. We measure @ 250 nm wavelength.
I started a sequence run with 1 standard at a concentration of 8 µg/ml of the highest target compound (RT 8 mins), and injected it to simulate a total analysis time of 71 hours. As you can see on the graph, the black chromatogram is the first injection and the red chromatogram is the last injection @ 71 hours.
The column supplier confirmed me that my LC-settings/parameters are acceptable for the column i use.
Increasing the temperature improved a bit the peak shape, but still not satisfactory (see graph). I was thinking to increase the equilibrating time to 7 mins instead of 5mins, but i wanted to hear what you guys were thinking about this. Can anyone help me to find a solution for this ?
i've troubles optimising our analytical method : the peaks of all the target compounds are broadening (peak fronting) in function of time (see the picture below).

We have an Agilent 1100 LC/MS (MWD + Ion Trap) and as you can see, i use a gradient system with following settings @ a flow rate of 0.4 ml/min:
00.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
08.00min 60% H2O / 40% Acetonitrile + 0.1% formic acid
08.50min 0% H2O / 100% Acetonitrile + 0.1% formic acid
09.50min 0% H2O / 100% Acetonitrile + 0.1% formic acid
10.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
15.00min 85% H2O / 15% Acetonitrile + 0.1% formic acid
The column is a 2.1 mm x 20 mm C18, temperature controlled @ 35.0 C and with an injection volume of 3 µl. We measure @ 250 nm wavelength.
I started a sequence run with 1 standard at a concentration of 8 µg/ml of the highest target compound (RT 8 mins), and injected it to simulate a total analysis time of 71 hours. As you can see on the graph, the black chromatogram is the first injection and the red chromatogram is the last injection @ 71 hours.
The column supplier confirmed me that my LC-settings/parameters are acceptable for the column i use.
Increasing the temperature improved a bit the peak shape, but still not satisfactory (see graph). I was thinking to increase the equilibrating time to 7 mins instead of 5mins, but i wanted to hear what you guys were thinking about this. Can anyone help me to find a solution for this ?