Now for the weirdness. I have two Alliance systems, 2695 Separations Modules with 2996 PDAs. One also has a 2487 DWD. They are both experiencing the same problem on the PDAs. That chromatogram is at 252 nm, but it looks the same at practically any wavelength. similar oscillation is seen on the DWD, but with a lower amplitude and a slightly different frequency at any given wavelength.
The problem started for all around the same time - digging back in the data, after a certain day all baselines look like above, but before that day they are perfect.
The pressure is extremely steady with a good delta (about 8 psi with a column attached) and does not track the oscillation. Same issue is seen whether a column or a union is is place, and with isocratic 100% methanol at 1 mL/min or any combination of gradient and solvent.
With a union in place, 100% methanol, 1 mL/min, pressure is 200 psi, delta 4 psi, and the chromatogram looks much like the one above. That particular chromatogram was acquired with a 100x4.6 mm 3.5 um C18 in place running a water/ACN gradient.
PDA diagnostics all pass. PM was completed by the vendor on Nov. 20 and was working fine for a while after the PM was completed. Reading the PDA calibration tool with 21 ms exposure time, I get a response of a little less than 10000 at 230 nm on the PDA for one instrument, which is a little low, but a very healthy response on the other PDA. However, both instruments have close to identical behavior.
Looks to me like an environmental factor affecting the detectors...
Temperature in the room is very stable. Problem started in the middle of winter so it doesn't seem like a change in the building's enviromental settings is a likely culprit, any major change should alreay be done. Instruments have been in the same exact location for a couple years with no problems.
The instruments are in a small room with several other LCs, which are not experiencing any problems.
Each instrument is on its own surge protector bar connected to the same wall plate which is in turn connected to a UPS. In the same room as several other LCs including two UPLCs which are experiencing no issues. I tried plugging the Alliance system into the same wall plate as the Acquities (different UPS) but no change. I tried swapping the power bars but no change.
I've considered the flow cells... perhaps something was run on the instruments at the same time which contaminated the flow cells, although I can find no evidence of it. I tried flushing with THF, and I executed a protocol from Waters to clean with 30% phosphoric acid, but no change in the issue.
I thought of whether there might have been a power surge, but the UPS logs don't record anything significant around that date.
We have a service contract, and Waters is sending me a new flow cell and lamp to try. But I'm increasingly skeptical as to what that's going to accomplish.
I wonder if anyone has ever seen anything like this before and can give me any ideas.