Signal Without turning on FID Detector!

Posted: Fri Oct 06, 2023 11:18 am
by ahmed ismail
Dear Colleagues,
apologizing for frequently wondering but I first faced this problem.
GC produces a signal without turning on the FID Detector!? Please look at the attached photo.
Image
Image
Image
Image

After turning on the detector and then conditioning the column while the instrument waiting for injection, the baseline appeared as in the attached image.
Image
for considering my type of sample is highly concentrated organic Solvents.
I assume that the carbon is scaled at the jet but I want to ensure from experts before I commence.

Re: Signal Without turning on FID Detector!

Posted: Wed Oct 18, 2023 3:02 pm
by James_Ball
If the detector is off it could simple be background noise in the electronics.

What is the scale of the signal?

If the absolute signal is in the single digits then just noise, if it is in hundreds or thousands then something may be shorting out in the detector.

Re: Signal Without turning on FID Detector!

Posted: Thu Oct 19, 2023 5:29 pm
by ahmed ismail
If the detector is off it could simple be background noise in the electronics.

What is the scale of the signal?

If the absolute signal is in the single digits then just noise, if it is in hundreds or thousands then something may be shorting out in the detector.
Yes, the value is small approximately 0.1 pA.

Re: Signal Without turning on FID Detector!

Posted: Mon Oct 23, 2023 6:55 am
by dblux_
If the detector is off it could simple be background noise in the electronics.

What is the scale of the signal?

If the absolute signal is in the single digits then just noise, if it is in hundreds or thousands then something may be shorting out in the detector.
Yes, the value is small approximately 0.1 pA.
What you guys mean by "detector is off" ?
No flame or electrometer switched off ?
How high is the signal when electrometer is off ?

Edit:
Just checked my FID. Signal is zero when the flame is off (no matter whether electrometer is on or off).

Re: Signal Without turning on FID Detector!

Posted: Mon Oct 23, 2023 4:13 pm
by ahmed ismail
If the detector is off it could simple be background noise in the electronics.

What is the scale of the signal?

If the absolute signal is in the single digits then just noise, if it is in hundreds or thousands then something may be shorting out in the detector.
Yes, the value is small approximately 0.1 pA.
What you guys mean by "detector is off" ?
No flame or electrometer switched off ?
How high is the signal when electrometer is off ?

Edit:
Just checked my FID. Signal is zero when the flame is off (no matter whether electrometer is on or off).
Dear, the detector is off, I mean the flame is off.

Re: Signal Without turning on FID Detector!

Posted: Tue Oct 24, 2023 9:06 am
by dblux_
It won't hurt if you thoroughly clean your detector. Plastic insulators should not be touched with bare fingers to avoid salt deposits.

BTW - I assume you have the signal in question with hot FID. When FID is cold small amount of moisture may cause elevated signal.