Cholestane Internal Standard Issues on GCMS

Posted: Wed Oct 04, 2023 3:16 pm
by egwatts
Hello!

I am trying to run sterols extracted from marine sediments on a GCMS. For our lab method, which has previously and successfully been used, we add a-cholestane as an instrument standard to normalize peak areas too. Recently, I noticed the cholestane peak areas vary widely between the calibration curve and samples even though the same amount is added to each vial. I don't think there is any cholestane in the samples because even injecting the same exact sample twice, the cholestane peak areas vary by 10x.

Thanks,
Emily

Re: Cholestane Internal Standard Issues on GCMS

Posted: Thu Oct 05, 2023 1:56 pm
by Consumer Products Guy
We used cholestane as internal standard for assay of cholesterol, but used GC-FID.

Make sure your GC inlet temperature is high-enough. And is the solvent composition of your standards and samples EXACTLY the same?

Re: Cholestane Internal Standard Issues on GCMS

Posted: Wed Oct 18, 2023 2:42 pm
by James_Ball
Also consider how often you do inlet maintenance. If the liner has not been changed recently it could be causing problems and also could need to trim a few inches from the front end of the column in case it has become contaminated.