We have a Thermo Trace 1300 GC with an ISQ QD mass spectrometer running Chromeleon 7.
The GC is set up in Gemini mode (i.e. dual injectors and columns for separate FID and MS samples). The instrument has stopped working because it is stuck trying to reach the initial oven temperature (40 oC). It will cool but stays at an elevated temperature (60 oC).
After opening the door, I can get the temperature to drop down down to 30 oC, so the sensor appears to be working. But after closing the door, the oven rapidly returns the temperature back to a stable 60 oC, which prevents the sample queue and injector sequence from starting. The status displays "Waiting for oven". After 15 minutes the sequence queue will time out.
However, if I run a gradient method with an elevated initial temperature (50 oC), the sequence will start and run properly, but it will not return all the way down past 60 oC.
I have checked all the methods to ensure there are no aberrant method parameters present. The software has been reset and instrument powered on/off several times without resolving the issue.
Any ideas ?
Any ideas ?