Very old HPLC. Need software to run it.
Posted: Mon Aug 28, 2023 7:04 pm
by LabTechieMike
I have an old HPLC donated to the bio lab at a local community college. I am responsible for getting it running. It is a Waters 1525 pump with a 717plus autosampler and a 2996 PDA. I was running on a workstation that had Empower 2 on it, but I do not have the login information for the admin on that workstation. I still have the PCI busLAC/E card. Does anyone know of a software solution that I can use? I contacted Waters and they told me it was too old and the busLAC/E card is no longer supported by their software. Any ideas?
Re: Very old HPLC. Need software to run it.
Posted: Mon Aug 28, 2023 8:27 pm
by Hollow
What admin account are you reffering to? The one on Windows or in Empower?
For the first, there are protocols and tools out there to reset windows passwords, if you can start the PC from USB or CD/DVD drive.
On Empower, if the default admin account was not disabled nor its password changed, you may want to have a look on the Empower software installation guide...
Posted: Mon Aug 28, 2023 8:47 pm
by LabTechieMike
I was referring to the admin account in Empower 2. I did try the default system login, it has been disabled as the previous users were GLP or close to GLP compliant. Thank you for the help.
Posted: Tue Aug 29, 2023 9:08 am
by Hollow
Do you have the installation media?
maybe you can talk Waters into deactivation of your license so you can do a fresh installation and reactivation?
Posted: Wed Aug 30, 2023 1:21 pm
by LabTechieMike
I have contacted Waters, they are looking into getting me the license that I am going to need to make the software run. Thank you
Posted: Thu Aug 31, 2023 1:38 am
by KarenA001
They will want you to pay for it as Waters licenses are not transferable...
- Karen
Posted: Fri Sep 01, 2023 2:32 pm
by DR
The latest version of Empower may not support IEEE, and they don't have replacement bus-lace boards anymore, but I suspect they can get you up and running with a supportable version of Empower 3 for a stand alone PC. You should even be able to run it with Windows 10. You may need a pump controller for that old pump...
You'd be running at risk of being dead in the water should the bus-lace board fail. There are used ones available (labX, ebay etc. etc).
As you're in education, they should be able to get you going for a somewhat discounted cost, and you'd save more by buying your own mini-tower PC and dropping a LTSC build of Windows 10 on it. Get one with a good (gaming) motherboard, plenty of RAM and a big SSD and it should last as long as you care to run the old system.
Good luck.
Posted: Fri Sep 01, 2023 5:46 pm
by LabTechieMike
You were close, I needed a Windows 7 machine, but waters took care of me and we are up and running. Thanks for the help