I have performed a sequence (isocratic RP method, acetonitrile/acetate buffer pH 5) where I get a negative peak in the blank at the exactly same retention time as the main peak in the samples/standards
The main peak has good retention at about 13 minutes (t0 is about 1 minute). My blank sample is just mobile phase which is also used to dissolve standards and samples.
Detection by UV at 383 nm
It doesn't make any sense? Do you have any idea what is going on?
Re: Negative peak in blank injection at RT of main peak
Posted: Tue Aug 22, 2023 6:56 pm
by tom jupille
My first guess would be that your mobile phase is contaminated with some of your product while the blank is "clean". In effect, the blank represents a "negative concentration" of analyte relative to the mobile phase.
Re: Negative peak in blank injection at RT of main peak
Posted: Thu Aug 24, 2023 7:16 am
by Mattias
Thanks Tom,
That makes sense, although I cannot understand how I managed to contaminate my eluent.
The analyte in question is Aniline Yellow and the concentration range in the samples is very low (few nanograms/ml). Maybe the vapour of the substance was enough to cause this. I will repeat the experiment and be more careful

Re: Negative peak in blank injection at RT of main peak
Posted: Thu Aug 24, 2023 5:19 pm
by vmu
I get a negative peak in the blank at the exactly same retention time as the main peak in the samples/standards
This may be just a coincidence, but may be not.
Detection by UV at 383 nm
Don't you use a DAD and (unnecessary, wrong) reference wavelength?