Pulsed Splitless injection of FPD

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I have noticed a coworkers method and see something that does not make sense to me. He inherited the method from previous employees. Using a 30M x 0.25mm x 0.25uF column the head pressure at idle is 44 PSI. The method has a setting for pulse pressure of 25 PSI and pulse time of 0.3 minutes. The purge time is set for 0.75 minutes. After the pulse time the method goes back up to 44 PSI head pressure.

I thought the point of pulse was to increase the pressure at injection time for a very short time to keep the sample from causing a back flash. This method does the exact opposite by lowering pressure for the splitless injection - and raising it for the remainder of the method. If there is any reason for doing this pressure drop on injection?

Thanks.
It looks like an error to me.
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry