Processing TPH C6-C9 Through LabSolutions/Shimadzu (GC-MS)

Posted: Tue Jul 25, 2023 8:34 am
by PurplePeacock
Hi All,

Some background: Our lab usually runs Volatiles TPH's on an agilent GCMS through headspace. This is then analysed through chemstation using a SCAN method that integrates from C6 through to C9 with one baseline.

The Problem: I am having an issue setting this up through Shimadzu GCMS (NEXIS GC-2030, GCMS-TQ8050 NX) and the LabSolutions software. The actual run is fine, all peaks are coming through nice and sharp.

I have been able to set up a program inside the qualitative parameters window, using the "PEAK DETECT", "PEAK TOP", "PEAK END". When i choose a chromatogram and simulate this I am getting the single baseline integrating from C6-C9 perfectly. BUT I am unsure how to make this program apply to a batch of calibration standards (we run 20 ppb to 2000 ppb), and give me a quantified result.

I am fairly new to building a process method from scratch, mainly just prep samples and process/analyse, so I apologise if I am missing any obvious information that would make this question easier to answer.

Any help would be greatly appreciated.

Kind regards,

Lochlan

Re: Processing TPH C6-C9 Through LabSolutions/Shimadzu (GC-MS)

Posted: Fri Jul 28, 2023 2:28 am
by PurplePeacock
Solved (by shimadzu): We ended up having the quantitative Scan method set up in Insight instead of lab solutions.