I was asked the following (rewritten so that it makes sense) by a student and was a little bit stumped on the best way to answer. Your thoughts would be most welcome!
“SIM or MRM methods tend to utilise a single defined quant ion/transition and one or more qualifier ions/transitions for each analyte. In most instances the quant ion is used to define integrated peak area (response) and the qual ions are only there to provide qualification that the peak is substance X; they don’t themselves normally contribute to the peak integration.
The software that we use for analysis allows response to be derived from the quant ion only or alternatively the quant ion plus the qualifier ions. The latter obviously provides greater absolute response values, so why do most analyses seemingly define peak response based upon the qualifier ion only and what are the issues with using all ions?”
My initial thoughts are that responses based upon a quant ion only are likely to have less variation than those derived from all ions, but this would depend fairly heavily on the nature of the qual ions. I did play around with a few batches of data to prove/disprove the point and there seems to be some truth in this idea, but RSD gains by using a single quant ion v using all ions seemed to be minimal although the data was not low level. At the same time, increases in integrated peak area (in the case of some analytes with more abundant qual ions) were significant when all ions were used to integrate. I also considered SNR as a possible reason, and although the baseline of low-level quant ions is noisy, calculated SNR calculated (Peak to Peak or RMS) didn’t seem to change whether using all ions or a single quant ion.
Any thoughts on this?