ALS - G4513A - 7693A Autoinjector
Currently moved to a 10m x 0.1mm x 0.1um Column on Hydrogen Carrier Gas for speed gain. Previously 30m x 0.25mm x 0.250um Column.
Samples that come into lab are concentrated, to prevent column overload I have attempted to move to a smaller injection volume. I have purchased Agilent Blue Line 0.5ul and 1ul syringes attempting to inject 0.1-0.5ul into the inlet. The split is currently at 200:1, with that being as high as i would like to allow for multiple instruments on the same gas generator. I have been having reproducibility issues at low inj volumes. I noticed whenever i manually dispensed the needle the sample stays as a droplet so i added a dwell time of 0.5 post injection which seemed to improve the chromatography but my RSD for my injections are all over the place. I am also using fast injection. Can someone help me?
ALS Parameters
Syringe Size - 1ul
Inj Vol - 0.1ul (Smallest i have attempted)
Solvent washes - 6Pre&6Post
Sample washes - 6
Sample pumps - 5
Dwell time (PreInj) - 0 min
Dwell time (PostInj) - 0.5 min
Injection Dispense Speed - 600ul/min
Viscosity Delay - 0sec