Aldehyde DNPH derivatives and GCMS
Posted: Thu Jul 06, 2023 1:49 pm
by mspera
Can 2,4-dinitrophenylhydrazone derivatives of aldehydes, particularly that of formaldehyde, be quantified using GCMS? We have a GCMS and are looking into analyzing raw materials and products for their formaldehyde content. I see numerous papers and test methods which specify HPLC. Are the DNPH derivatives unstable to high temperatures? I know they can be used for melting points, but will they survive a 275 C GC inlet?
Re: Aldehyde DNPH derivatives and GCMS
Posted: Thu Jul 06, 2023 3:21 pm
by James_Ball
The other question would be if DNPH has a low enough boiling/melting point to be used in GC. It may not be volatile enough.
EPA Method 556.1 uses PFBHA 0-(2,3,5,6-PENTAFLUOROBENZYL)-HYDROXYLAMINE HYDROCHLORIDE as a derivitizing agent for GC/FID which should work for GC/MS also.
Posted: Thu Jul 06, 2023 8:17 pm
by rb6banjo
You will likely struggle with 2,4-DNPH derivatives and GC - even for formaldehyde.
I've used this and solid-phase microextraction to get H2CO:
https://www.sigmaaldrich.com/US/en/prod ... ich/194484
The idea was to load up a fiber (CX/PDMS) with reagent by exposing a fiber to the headspace of an aqueous solution of the reagent. Then, remove it and expose it to the sample for X amount of time. The derivatives are amenable to GCMS. I just used a flame but you could use MS. The trouble I had was that you can't get a tremendous amount of the reagent on a small SPME fiber so I had to expose the sample to the loaded fiber for very brief times (15-20 seconds). Otherwise, I risked running out of derivatizing agent on the fiber. In the end, I was able to determine H2CO in plywood with that procedure.
Posted: Fri Jul 07, 2023 1:52 am
by soccerdad007
I am not sure if you have the capability to do headspace but we used this method in past with reliable results
https://resources.perkinelmer.com/lab-s ... dehyde.pdf
Kevin
Posted: Tue Jul 11, 2023 12:47 pm
by mspera
Thank you all for your helpful responses. I will need to investigate this further.
Posted: Tue Jul 09, 2024 4:26 pm
by csiquet_flex2000
Hello everybody,
For my very first post, I will be commenting on this one year old topic as it is very similar to what I'm trying to do.
I managed to come up with a GC-FID method to separate and detect DNPH-propionaldehyde derivatives and I was trying to devise a quantification method.
My problem is that, according to this article, aldehydes (except formaldehyde) and asymmetric ketones will produce two isomers when derivatized and the ratio of the cis/trans isomers might be light dependent.
It would not be a problem if the response factors of the isomers were the same but they should be different.
So far, it seems the ratio has been mainly the same. I even injected a solution before and after sun exposure. The ratio stayed the same but the signal was lower (this suggested the solution is not stable to light).
Would you have a suggestion on how to procede with a quantification method?
I usually do external standard calibration curves with added internal standard but I'm really puzzled on how to consider the possibility of having a variable isomers ratio.
Thank you.
Posted: Tue Jul 09, 2024 11:18 pm
by Consumer Products Guy
We investigated this as well, and our conclusion was that DNPH derivatives of formaldehyde and glutaraldehyde were best assayed by HPLC.
Posted: Thu Jul 11, 2024 8:42 am
by csiquet_flex2000
Thank you for the answer.
Indeed, according to the article I mentioned, HPLC is better suited for this analysis. If the correct wavelength is selected because the isomers have different absorption spectra.
The thing is, the peak for DNPH-propionaldehyde is not too well resolved from the matrix with our LC method. Further development is needed I guess.
Alternatively, I could use GC-MS I believe.
The difference of response factors seen in GC-FID should not be a problem in SIM mode.
Am I wrong?
Posted: Thu Jul 11, 2024 2:35 pm
by csiquet_flex2000
Just leaving this here for reference:
GC-FID - Application News from Shimadzu (NÂș G327A)